Welch 1400 flushing run

Posted: Sat Mar 17, 2018 12:13 pm
by Filippo_Passeri
Hello, all.

Today I finally acquired motor pulley for my Welch 1400 pump. The pulley was 3D printed since I could not find a replacement online. After it was put together, it's inlet was capped off and the pump was turned on for 45 minutes. Then the dirty, but metal particle free, oil was drained from the pump head. The pump head was then refilled with fresh oil and its performance was measured with a Mastercool thermocouple gauge. After two minutes the gauge read around 150 microns. I'm sure it deserves at least, a second flushing run.

During these trials, the sound the pump made was at first a bubbling noise that then receeded to a clicking noise similar to turning back a bicycle pedal.

Hope I'm on a path to a great pump for fusor and other vacuum-based experiments.
The whole pump during flushing, you can see the blue plastic pulley on the left of the picture.
The whole pump during flushing, you can see the blue plastic pulley on the left of the picture.

Re: Welch 1400 flushing run

Posted: Mon Mar 19, 2018 2:05 am
by Richard Hull
Yes, 150 microns is really bad. 1400's tend to come surplus in fairly worn out shape. Try another flush, as you note. I would be amazed at it hitting even 100 microns. I have had two surplus 1400's in my collect over the years and the very best the finest one ever did was 50 microns. Good luck.

Richard Hull

Re: Welch 1400 flushing run

Posted: Mon Mar 19, 2018 6:30 am
by Filippo_Passeri
It’s also possible, that the connection between the thermocouple and the pump inlet is not 100% tight, since the connection is not permanent, it has been sealed with teflon tape. I know it is not the best sealant, but its better than nothing. After the pump was turned of, the pressure rose drastically.

I’m still bit of a novice on sealing techniques. I know i can use locktite on threaded connections, but that’s a permanent connection. The connection is a NPT thread. I’m open to suggestions for improvements.

Re: Welch 1400 flushing run

Posted: Tue Mar 20, 2018 10:55 am
by Richard Hull
Turning the pump off without immediate venting of the line between the pump and gauge can easily draw pump fluid up into you gauge, ruining, or at best, fouling it! You should not rely on an unknown, used, surplus pump's anti-suckback valve. Some pumps do not even have them and on those that do, they maybe fouled or totally inoperable from horrid maintenance in their dark past.

Richard Hull