Foreline pump base pressure degraded

Every fusor and fusion system seems to need a vacuum. This area is for detailed discussion of vacuum systems, materials, gauging, etc. related to fusor or fusion research.
Post Reply
ian_krase
Posts: 636
Joined: Mon Nov 28, 2016 2:48 am
Real name: Ian Krase

Foreline pump base pressure degraded

Post by ian_krase »

Hello, everyone.

I use a Pfeiffer Duo (the newer version with the white case) as a foreline pump. When I first got this (used) pump, it performed pretty well with about 8 microns of base pressure. Now, however, I can't get lower than about 60 microns. I didn't really notice since lately I've been using it exclusively as a foreline pump.

The oil is fairly new though maybe I should replace it. The pump doesn't sound any different or anything like that.

Do you think I need to do an outright rebuild?
Bruce Meagher
Posts: 148
Joined: Thu May 12, 2011 11:25 pm
Real name: Bruce Meagher
Location: San Diego

Re: Foreline pump base pressure degraded

Post by Bruce Meagher »

Ian,

After helping Tyler through his pump rebuild it’s not evident the parts in the minor rebuild will have much impact on performance. There are a couple o-rings inside that could be leaking, but I'd think this would be unlikely. Of course cleaning during the rebuild process could make a huge impact if contamination is an issue. The first thing I always assume when I have a vacuum issue is that I have a leak somewhere. Have you taken the pump off the foreline and just tested directly at the intake port? Next would be contamination in the oil. How long ago did you measure the 8 microns (weeks, months, years,…). How much usage has the pump seen in that time, and what type of processes / contamination have been put through it? I probably put less than 100 hours a year on my vacuum pumps. Maybe it’s just water vapor and a few hour run with the ballast open would clear everything? Also, could it be a faulty or leaky gauge? I like to check / cross calibrate my gauge with another system to make sure they are in good working order. A couple pictures of your test setup might be helpful.

Bruce
ian_krase
Posts: 636
Joined: Mon Nov 28, 2016 2:48 am
Real name: Ian Krase

Re: Foreline pump base pressure degraded

Post by ian_krase »

I tested with nothing but a valve between the pump and the gauge, should go with nothing but the gauge.

8 microns was measured over a year ago, with the pump on a system reaching sub-30-microns routinely for months later. However, I haven't measured the pump's ultimate vacuum in some months since then until now.

By rebuild I meant a more complete rebuild including replacing the vanes. Have found a source, but don't want to do it if not needed -- 70 dollars plus hours of disassembly on grease-spattered table.

Really pretty much just air that has gone through it though it's been exposed to trace quantities of solvent vapors and plasma-burned dust-off occasionally. Total of a few hours run a week.

The gauge is unlikely to be faulty or leaky -- it's a new-to-me convectron that I tested the calibration of at both zero and atmosphere, though it always could have a cracked feedthrough or something.
User avatar
Richard Hull
Moderator
Posts: 14992
Joined: Fri Jun 15, 2001 9:44 am
Real name: Richard Hull

Re: Foreline pump base pressure degraded

Post by Richard Hull »

The obvious assumption is that you have fresh oil in the pump of very recent input. If not and you want to try to keep the oil. Follow the suggestion of gas ballasting for two to three hours with the inlet blocked off. Smart money would just replace the oil. with new oil, still gas ballast for a few minutes before attaching the TC gauge tube. Glow cleaning and blasting off dust over a year will foul oil.

Richard Hull
Progress may have been a good thing once, but it just went on too long. - Yogi Berra
Fusion is the energy of the future....and it always will be
The more complex the idea put forward by the poor amateur, the more likely it will never see embodiment
ian_krase
Posts: 636
Joined: Mon Nov 28, 2016 2:48 am
Real name: Ian Krase

Re: Foreline pump base pressure degraded

Post by ian_krase »

I tried blankoff and gas ballast for a while but the only result was a bit of foaminess in the oil. Going to try changing the oil when I have a chance.
User avatar
Richard Hull
Moderator
Posts: 14992
Joined: Fri Jun 15, 2001 9:44 am
Real name: Richard Hull

Re: Foreline pump base pressure degraded

Post by Richard Hull »

Foam in the oil is a definite sign of air and moisture in the oil. OOPs ....something is not sealed well.

Richard Hull
Progress may have been a good thing once, but it just went on too long. - Yogi Berra
Fusion is the energy of the future....and it always will be
The more complex the idea put forward by the poor amateur, the more likely it will never see embodiment
Mark Kimball
Posts: 20
Joined: Sat Jun 03, 2017 12:26 am
Real name: Mark Kimball

Re: Foreline pump base pressure degraded

Post by Mark Kimball »

At work we had a nice gold sputtering system for preparing samples for SEM examination. It used a small 2-stage mechanical pump to evacuate the chamber. It didn't take long for the pump oil to absorb a _lot_ of water (this was in Oregon, infamous for its damp weather). Symptoms included extended pump-down time and foamy pump oil observed in the pump's sight glass. You might be able to enable the ballast on your pump and -- slowly -- pull the water out of the pump oil. A CDA or dry nitrogen purge would be best, if you have access to either one.

We also had an RIE (reactive ion etcher) that used a turbo pump + mechanical pump to evacuate the etch chamber. Most who used it would leave the thing at atmospheric pressure after they were done. Very few understood (or perhaps cared) that the next user would have to wait much longer to get the chamber down to operating pressure as the water (slowly) desorbed from the chamber walls.

Mark
ian_krase
Posts: 636
Joined: Mon Nov 28, 2016 2:48 am
Real name: Ian Krase

Re: Foreline pump base pressure degraded

Post by ian_krase »

Oh boy. I've been letting my whole system up to atmosphere this whole time.

New oil it is.
User avatar
Richard Hull
Moderator
Posts: 14992
Joined: Fri Jun 15, 2001 9:44 am
Real name: Richard Hull

Re: Foreline pump base pressure degraded

Post by Richard Hull »

Your whole system??? I thought this was just a blank off test. You can let a pump, by itself, up to air on a regular basis. When you do this, it is wise to cap off the inlet so constant exposure to atmosphere is obviated.

Entire systems, plumbing, diff pumps, fusor chamber, etc., are best left under a valved off vacuum and never let up to full atmosphere unless repairs of major components are needed.

Letting a mechanical fore pump up to air is normal provided its inlet is then sealed off from contact with outside air over periods of time.

Richard Hull
Progress may have been a good thing once, but it just went on too long. - Yogi Berra
Fusion is the energy of the future....and it always will be
The more complex the idea put forward by the poor amateur, the more likely it will never see embodiment
Tom McCarthy
Posts: 405
Joined: Wed May 08, 2013 3:36 pm
Real name: Tom McCarthy
Location: Ireland
Contact:

Re: Foreline pump base pressure degraded

Post by Tom McCarthy »

The first part of your message gave me a shock Richard, relieved to see the caveat for forepumps.

My standard procedure when finished pumping and using the system is to first turn off the turbo pump. I leave the mechanical forepump running. After an hour or so I look for a noticeable rise in pressure as a proxy for turbo slowdown. Then I close the valve between rotary pump inlet and turbo outlet and turn off the turbo pump. Finally I undo the KF connection on the forepump inlet and bring the inlet up to air, then re-clamping the connection.
User avatar
Richard Hull
Moderator
Posts: 14992
Joined: Fri Jun 15, 2001 9:44 am
Real name: Richard Hull

Re: Foreline pump base pressure degraded

Post by Richard Hull »

You method is good and how I do it. I valve off the fusor chamber to diff pump inlet. (fusor left under vacuum. I then turn off the boiler in the diff pump and monitor the diff pump boiler temp until it drops to below 80 deg C, then I valve off the diff pump outlet to the mechanical pump inlet. (diff pump left isolated and under vacuum. Last I turn off the mechanical rotary pump and immediately un-clamp the hose to the mechanical pump and foreline. I let the foreline and mechanical pump up to air for 1 second (woosh!) then immediately re-clamp the foreline to pump.

I also have a ball valve on the mechanical pump's outlet. I immediately close this valve so that no air can enter the turned off rotary mechanical pump until the next time I use it.

Naturally, the first action on any restart is to open this ball valve to allow the rotary to exhaust properly while running.

Richard Hull
Progress may have been a good thing once, but it just went on too long. - Yogi Berra
Fusion is the energy of the future....and it always will be
The more complex the idea put forward by the poor amateur, the more likely it will never see embodiment