He3 corona tube + oscillocope

This area is for discussions involving any fusion related radiation metrology issues. Neutrons are the key signature of fusion, but other radiations are of interest to the amateur fusioneer as well.
Post Reply
RealBorg
Posts: 27
Joined: Thu Mar 01, 2012 3:40 am
Real name: Thomas Zehetbauer
Contact:

He3 corona tube + oscillocope

Post by RealBorg »

hi there,

I want to test my He3 tube (CHM32) by hooking it up to a power supply limited to 2400V (diode in reverse) and 25µA (10MOhm resistor) and wait for background impulses.

How should I connect the oscilloscope?
Would a simple voltage divider probe be sensitive enough?
Should I connect a capacitor and limit voltage to the oscilloscope with a diode?
Is it better to measure Delta-I by connecting the tube with inductive coupling to the oscilloscope?

TIA
Tom
User avatar
Richard Hull
Moderator
Posts: 14992
Joined: Fri Jun 15, 2001 9:44 am
Real name: Richard Hull

Re: He3 corona tube + oscillocope

Post by Richard Hull »

Either method should work, but simple capacitive coupling is hard to beat. Use a .001ufd 3 or 4 kv capacitor hooked to the tube's central wire and connect the scope through that. Of course, use the limiting resistor from the supply to the central wire to limit the current. Bring the supply voltage up slowly until you see pulses. You should put the scope on or near its lowest voltage setting.

I hope your scope is a storage type to catch the first pulse. Once captured you can adjust the scope to display the full pulse. reset to another trigger and capture and keep adjusting the scope until you have a nice pulse on the screen. This will tell you a lot about what kind of pre-amp your will need.

Richard Hull
Progress may have been a good thing once, but it just went on too long. - Yogi Berra
Fusion is the energy of the future....and it always will be
The more complex the idea put forward by the poor amateur, the more likely it will never see embodiment
Post Reply

Return to “Neutrons, Radiation, and Detection (& FAQs)”