Second pulse from He3 tube
Lukas Springer
Real name: Lukas Springer
Post by Lukas Springer »
Hi!
I finally got myself a Helium-3 corona tube, the SNM-18-1 to be precise.
I don't have a neutron source yet, so I've got to rely on background for testing; the screenshot below shows the signal from my tube for a couple of minutes, without a moderator.
The tube counts about 10 cpm, and some (but not all) of these pulses are followed by a second one with about 650 us delay.
I don't understand where that pulse comes from, my only idea is a recombination event. Maybe this it is from UV ionisation when the proton "catches" an electron?
-Lukas
He3 tube second peak
Re: Second pulse from He3 tube
Make sure the instrument cables are terminated by their characteristic impedance. This is usually a BNC connector with a built in resistor. The NIM system uses 94 ohms, but both 50 ohms and 75 ohm cables exist in other areas of high frequency electronics.
Andrew Seltzman
Real name: Andrew Seltzman
Post by Andrew Seltzman »
In this case I don't think termination or ringing of a cable is the issue; a 650us delay would equate to a 400000ft long RG58 cable (5 ns/m (1.54ns/ft) delay).
Andrew Seltzman
www.rtftechnologies.org
Lukas Springer
I don't think it's from bad / missing termination, since not every pulse is followed by a second one, regardless of pulse height.
There is currently no amplifier installed and I'm measuring with the standard 1M input impedance of my scope, right after the AC coupling capacitor.
All cables are 1m in length and everything is shielded, the passives are in a soldered shut box.
I would not be too excited about the second pulse if this has no amp or discriminator. This is not a real neutron detector yet. Get some electronics in there and see what you really have.
Lukas Springer
Of course it isn't a detector yet, it's just testing the tube so I could make use of ebays return policy, if needed.
The scope is placed where the amp would be, the amp will have 1M input impedance as well, so it should amplify that smaller, second pulse.
I know this pulse is not a neutron count, I'm just trying to understand where it comes from.
Once the setup is complete (I've cast the paraffin moderator today) I'm going to ask my university for a californium check source, then I can tell you if it is in any way related to a neutron event.
