Ludlum 42-31

Posted: Sun May 12, 2019 3:27 pm
by Joe Gayo
I'm trying to figure out the correct operating voltage for this tube (Ludlum 42-31).

The website for Model 42-31H Neutron Detector says 1200V but the LND referenced by the same page says 750V (https://www.lndinc.com/products/neutron ... ors/25185/).

It's also very possible that the "H" means He3 and the non-H variant is a BF3 tube. The BF3 tube referenced 01-5018 (https://ludlums.com/images/pdf_files/Mo ... f_tube.pdf) is said to be a Nancy Wood G-5-1 (https://www.qsl.net/k/k0ff/0%2018Manual ... 0jun98.pdf) that operates at 1600-1900V.

Through a window on the tube it says 201 (could be partial) and this lines up to a couple of LND BF3 tubes but nothing exactly.

Re: Ludlum 42-31

Posted: Mon May 13, 2019 12:55 am
by Richard Hull
I would work my way up in voltage starting at 750v. You have a great working fusor and you will see neutrons aplenty as you raise the voltage, slowly.
This assumes you never get a firm piece of data on the tube. BF3 tubes tend to take a lot more bias than 3He tubes, Boron lined detectors are more in the 600-800 volt range and are very critical of voltage adjustment.

Richard Hull

Re: Ludlum 42-31

Posted: Tue May 14, 2019 3:51 pm
by Bob Reite
BF3 tubes are usually operated between 1800 and 2000 volts. The "sweet spot" for mine is at 1900 volts.

Re: Ludlum 42-31

Posted: Tue May 14, 2019 8:06 pm
by Joe Gayo
I see some reasonable response at 1100 - 1200 V but I have the secondary amplifier gain set significantly higher (128x versus 4x) than my other detection system. I suspect that I need to raise the voltage but I'm reluctant since I don't want to damage my tube.

Re: Ludlum 42-31

Posted: Wed May 15, 2019 10:54 pm
by Bob Reite
If indeed, you have the 42-31 3He tube, that should be operated no higher than 1100 volts. But if your model number is just "42-31" that would be the BF3 tube which is rated at 2000 volts max.

Re: Ludlum 42-31

Posted: Fri May 17, 2019 1:05 am
by Joe Gayo
Bob,

Just so I'm clear you have a 42-31 and bias it at 1900V, or you are talking about BF3 tubes in general.

Thanks, Joe

Re: Ludlum 42-31

Posted: Sat May 18, 2019 12:25 am
by Bob Reite
BF3 tubes in general operate between 1800 and 2000

Re: Ludlum 42-31

Posted: Sat May 18, 2019 1:06 pm
by Bruce Meagher
One way to determine if you have a BF3 or He3 tube is to look at the output pulse height spectrum. The two spectrums are significantly different. Knollā€™s book covers this in detail, but there are many practical examples in past fusor posts. One He3 pulse height spectrum I captured is described in this thread:

viewtopic.php?t=10569

Here are some other useful ones:

viewtopic.php?t=5613
viewtopic.php?t=6308

With your fusor's outstanding neutron production the features of these different spectrums should be easily identifiable.

Bruce

Re: Ludlum 42-31

Posted: Sat May 18, 2019 8:49 pm
by Joe Gayo
I definitely have a BF3 tube and the sweet spot is 1850V. Thank you, everyone, for your help.

P.S. I now have 2 NIM electronic detection systems, 1 portable electronic system, and a bubble dosimeter.

Re: Ludlum 42-31

Posted: Sun May 19, 2019 11:28 pm
by Bob Reite
Hopefully you are ready to make neutrons soon, as the bubble detectors have a limited shelf life, used or not. Use the bubble detector to calibrate your electronic detectors.

Re: Ludlum 42-31

Posted: Sun May 19, 2019 11:56 pm
by Richard Hull
Bob is correct. Assuming you have a good, new, bubble detector, use its calibrated dosimeter calculation (bubs/mrem) to calibrate your electronics.
I created a FAQ to explain how I did it years ago.

Richard Hull