Re: I think I have a fusor chamber.

Post by Dennis P Brown » Sun Nov 12, 2017 5:08 pm

Controlling the current to the diffusion pump heater sounds like a waste of time and effort. Controlling the vacuum rate via a gate valve is extremely fast; controlling current to a heater takes a long time (even for major current changes.)

My main gate valve is a manual butterfly type and works extremely well.

Silviu Tamasdan
Posts: 113
Joined: Wed Jun 28, 2006 11:17 pm
Real name: Silviu Tamasdan

Re: I think I have a fusor chamber.

Post by Silviu Tamasdan » Sun Nov 12, 2017 7:10 pm

Eh, it's not complicated to do (I have a PWM already built with the appropriate rating) and at worst it won't be useful. I was thinking that it may be handy for long runs at reduced pumping speed (not to make sudden changes during a run). Especially since it looks like I'm leaning more towards using a pump that is oversized for my chamber's needs.
There _is_ madness to my method.

User avatar
Richard Hull
Site Admin
Posts: 10473
Joined: Fri Jun 15, 2001 1:44 pm
Real name: Richard Hull

Re: I think I have a fusor chamber.

Post by Richard Hull » Mon Nov 13, 2017 7:13 am

Throttling the boiler electrically sounds bizarre. I would think the oil is either hot enough to form the needed jetting in the stack or it is not. This is what I observe at pretty much a single temperature level in my diff pump. It starts working at a specific temperature, dropping the fusor's pressure to a given sub micron level and a temperature climb of another 40 deg C does not deepen the vacuum one iota. It is kind of an on-off senario. One moment It's not pumping at all then, suddenly, it is going gang busters. Diff pumps do not seem to incrementally pump with increasing temperatures.

Once you actually run a diff pump with a pressure gauge capable of seeing it start to pump, you will observe this. Once it starts to work that is as low as it will go. There is no middle ground of variable pumping speed. The throttle valve on top of the diff pump to the fusor controls and can vary the pumping speed.

Richard Hull
Progress may have been a good thing once, but it just went on too long. - Yogi Berra
Fusion is the energy of the future....and it always will be
Retired now...Doing only what I want and not what I should...every day is a saturday.

John Futter
Posts: 1211
Joined: Thu Apr 22, 2004 2:29 am
Real name:
Contact:

Re: I think I have a fusor chamber.

Post by John Futter » Mon Nov 13, 2017 9:05 am

Diff pumps are a semi closed loop feedback system
DC 704 will run at one temp with a standard heater DC705 at another temp albeit higher
If you use Santovac it will be an even higher temp.
The diff pump is a self regulating system using the cooling in the upper parts and the super heated fluid in the lower parts ( giving the vapour flow) to establish the correct operating conditions all of this due to pump design and operating oil characteristics.
do not try to circumvent this setup it will lead to tears!!!
a burping unstable pump is not another variable you need to introduce to a fusor setup.
side effects will include contaminated vacuum gauges and chambers from this burping (farting) phenomeom and silicon oil coated surfaces are a B#$%^tch to clean properly.
Been there
Done that
Got the Tee shirt

Silviu Tamasdan
Posts: 113
Joined: Wed Jun 28, 2006 11:17 pm
Real name: Silviu Tamasdan

Re: I think I have a fusor chamber.

Post by Silviu Tamasdan » Mon Nov 13, 2017 11:22 am

OK scratch that part of the plan. Everything else stays the same though. The HV feedthrough definitely needs to go into one pf the KF50 because that is the only way to adapt mine.
There _is_ madness to my method.