Re: Joe Gayo's Fusor - Neutron Detection Upgrade

Post by Richard Hull »

My First fusion was with a mechanical pump that could hit 5 microns, backed by a micromaze that bottomed my TC gauge. (actual vacuum unknown). The best I ever did with fusor III was 250kn/s but I never got above 32 kv. Now that we are into the wall loading concept, day to day operation will see a bit of improvement. In addition, when fusor III was in operation I have far less knowledge and much reduced neutron detection (Eberline PNC-1) a good counter, but with a small Nancy Wood BF3 tube in it.

A word of advice. When you have your bubble detector fix forever your neutron tube and moderator or mark its position well. Calibrate your counter using the bubble detector. over many days and many runs recording all data counts during a run and the number of bubbles per run. Do this as many times as you can over the life of the bubble detector. I have many, many pages of such data. In the end, Based on the bubble detector's rmem dose back figure to the digital counters run total counts, you will arrive at a multiplication factor for future total counts that you can use in future long after the bubble detector is dead to arrive at a total isotropic emission per second of neutrons from your fusor. The accuracy of the multiplier will rely highly on the number of data samples taken over time. More samples, obviously the better the statistics and the resultant multiplier.

My multiplier turned out to be 21.6. I backed it off to an easy and conservative 20.0.

Richard Hull
Progress may have been a good thing once, but it just went on too long. - Yogi Berra
Fusion is the energy of the future....and it always will be
The more complex the idea put forward by the poor amateur, the more likely it will never see embodiment
User avatar
Joe Gayo
Posts: 404
Joined: Sun Jan 06, 2019 9:34 pm
Real name: Joe Gayo
Location: USA

Re: Joe Gayo's Fusor - Neutron Detection Upgrade

Post by Joe Gayo »

Will do with the bubble detector!

I assume when you say micromaze you are referring to a sorption pump?
User avatar
Joe Gayo
Posts: 404
Joined: Sun Jan 06, 2019 9:34 pm
Real name: Joe Gayo
Location: USA

Re: Joe Gayo's Fusor - Neutron Detection Upgrade

Post by Joe Gayo »

The following is for those who are looking for the lowest financial bar for a neutron producing fusor:

If you have reasonable strong electronics knowledge, then a sub $1000 fusor is possible maybe even sub $500.

** Use a high quality two-stage rotary vacuum pump (Edwards, Pfiffer, Precision Vacuum, etc.) - no diffusion or turbo pump required. A base pressure bellow 7mTorr seems sufficient. This also means you only need a thermocouple vacuum gauge.

** KF 40 cross makes an excellent chamber. KF fittings are cheap and simple to assemble / disassemble.

** Purchase NIM BIN equipment even if it's not guaranteed to work. Schematics are readily available in the manuals. The common failure points are cold solder joints, failed output BJTs, and tantalum capacitors. (I fixed/refurbished my NIMs and preamp)

** Purchase the OEM models of Spellman or Glassman power supplies, or better yet, units with just the multiplier stack (viewtopic.php?f=11&t=12625)

** Spark plug high voltage feedthrough - https://www.ebay.com/itm/Metric-M14-x-1 ... 2480729252 with a Champion J99 599 industrial spark plug. As long as the air side is sufficiently insulated you should be able to reack 20kV+.

My fusor was around $720.
User avatar
Richard Hull
Moderator
Posts: 14991
Joined: Fri Jun 15, 2001 9:44 am
Real name: Richard Hull

Re: Joe Gayo's Fusor - Neutron Detection Upgrade

Post by Richard Hull »

The micromase was pushed by Lesker for a few years and is no longer available. It was great and could warrant about one order of magnitude below a good mechanical pump that could go under 10 microns on its own. It had its own internal heater to purge bad stuff out and clean itself up. It was found by me that a full one hour pumping, heated purge of the micromaze and gas ballasting for each days run of the fusor. This hassle might have been the complaints around their micromaze and its demise. I did not mind it in those early days.

Richard Hull
Progress may have been a good thing once, but it just went on too long. - Yogi Berra
Fusion is the energy of the future....and it always will be
The more complex the idea put forward by the poor amateur, the more likely it will never see embodiment
User avatar
Joe Gayo
Posts: 404
Joined: Sun Jan 06, 2019 9:34 pm
Real name: Joe Gayo
Location: USA

Re: Joe Gayo's Fusor - Neutron Detection Upgrade

Post by Joe Gayo »

Bubble detectors arrived (55 and 54 bub/mrem)!

Did a quick 30 second run at near full power 45kV, 15mA, and 28mTorr.
Dosimeter has a holder that places it 22.4 cm from the center of the chamber.

5 bubbles ... 545,000 n/s TIER
IMG_1050_sm.jpg
Obviously I have more runs planned, but I couldn't wait to post.
User avatar
Dennis P Brown
Posts: 3159
Joined: Sun May 20, 2012 10:46 am
Real name: Dennis Brown

Re: Joe Gayo's Fusor - Neutron Detection Upgrade

Post by Dennis P Brown »

Very nice results; Bubble detectors are useful and the gold standard.

I too checked mine with a 22 bubbles/rem - turns out my neutron detector wasn't too far off (around 170 k neutrons/sec.) I did two chamber sizes with two bubble measurements to show that halving the fusor volume nearly doubled the neutron count (gas pressure and voltage all the same.) Was hoping someone else would look into that question and do more definitive experiments - one experimental run is not exactly proof.
Cristiano_Machado
Posts: 39
Joined: Fri Aug 24, 2018 9:03 am
Real name: Cristiano Machado
Location: São Paulo - Brazil

Re: Joe Gayo's Fusor - Neutron Detection Upgrade

Post by Cristiano_Machado »

Hello Joe,

Nice work and I am happy the BD have arrived earlier than expected.

I am trying to send you a private massage about the BD, but I don't know why the messages can't go out of my outbox. Can you contact me about our deal? Fell free to email me directly (c_machado@yahoo.com).

Regards,

Cristiano
Pablo Llaguno
Posts: 104
Joined: Sun Feb 05, 2017 6:00 pm
Real name: Pablo Llaguno

Re: Joe Gayo's Fusor - Neutron Detection Upgrade

Post by Pablo Llaguno »

Great work Joe!

I am curious as to how you set up your vacuum and D2 system. Does your setup use any valve between the pump and chamber? If not, how do you manage to not waste so much D2?
User avatar
Joe Gayo
Posts: 404
Joined: Sun Jan 06, 2019 9:34 pm
Real name: Joe Gayo
Location: USA

Re: Joe Gayo's Fusor - Neutron Detection Upgrade

Post by Joe Gayo »

I do have an isolation valve but I do flow the D2 and waste a decent amount at this point. My runs are fairly short.
User avatar
Joe Gayo
Posts: 404
Joined: Sun Jan 06, 2019 9:34 pm
Real name: Joe Gayo
Location: USA

Re: Joe Gayo's Fusor - Neutron Detection Upgrade

Post by Joe Gayo »

Good statistical run...

98 bubbles in 180 seconds, 12.5 cm distance

554,000 n/sec TIER

45kV 16mA

28mTorr
42B17938-41F7-44CF-91FD-AB00978F44F7.jpeg
User avatar
Joe Gayo
Posts: 404
Joined: Sun Jan 06, 2019 9:34 pm
Real name: Joe Gayo
Location: USA

Re: Joe Gayo's Fusor - Neutron Detection Upgrade

Post by Joe Gayo »

IMG_1071_sm.jpg
23Jan19-8am-Run.png
User avatar
Dennis P Brown
Posts: 3159
Joined: Sun May 20, 2012 10:46 am
Real name: Dennis Brown

Re: Joe Gayo's Fusor - Neutron Detection Upgrade

Post by Dennis P Brown »

Outstanding data processing and presentation. Perfect example of a top level experiment.

If you had done the IEEE paper, it would have been done correctly with carefully measured results.
User avatar
Joe Gayo
Posts: 404
Joined: Sun Jan 06, 2019 9:34 pm
Real name: Joe Gayo
Location: USA

Re: Joe Gayo's Fusor - Neutron Detection Upgrade

Post by Joe Gayo »

Hovering right around the 1 Mega n/sec mark ... 950k to 1050k per second ... once I upgrade my feed through and push well past 50kV the barrier should be definitively broken
User avatar
Richard Hull
Moderator
Posts: 14991
Joined: Fri Jun 15, 2001 9:44 am
Real name: Richard Hull

Re: Joe Gayo's Fusor - Neutron Detection Upgrade

Post by Richard Hull »

Silver will activate easily for you at you current level.

Richard Hull
Progress may have been a good thing once, but it just went on too long. - Yogi Berra
Fusion is the energy of the future....and it always will be
The more complex the idea put forward by the poor amateur, the more likely it will never see embodiment
User avatar
Joe Gayo
Posts: 404
Joined: Sun Jan 06, 2019 9:34 pm
Real name: Joe Gayo
Location: USA

Re: Joe Gayo's Fusor - Neutron Detection Upgrade

Post by Joe Gayo »

Correct, Silver is easy for my setup. I’m finishing my gamma spectrometer to work with Indium.
User avatar
Dennis P Brown
Posts: 3159
Joined: Sun May 20, 2012 10:46 am
Real name: Dennis Brown

Re: Joe Gayo's Fusor - Neutron Detection Upgrade

Post by Dennis P Brown »

Amazing setup and really great posts! Your efforts have been really impressive.

One thing you might consider adding is paraffin shielding (and possibly a boron compound added to its outer surface) since you are getting well over (another post) a million neutrons/sec. For longer term operation, this would reduce exposure - while your fusor does not produce a lot of radiation the neutrons are deeply penetrating and have a high impact on tissue compared to most other types of radiation (esp. due to these particles being fast neutrons.) Neutrons are, unlike other forms of radiation, rather long lived (they have a 10 min. half life) and do tend to "hang around" when they are thermalized.
User avatar
Joe Gayo
Posts: 404
Joined: Sun Jan 06, 2019 9:34 pm
Real name: Joe Gayo
Location: USA

Re: Joe Gayo's Fusor - Neutron Detection Upgrade

Post by Joe Gayo »

Added an axial magnetic field (~250 Gauss) to stabilize the discharge and push the power higher.

Pressure: 28mTorr-31mTorr (because the chamber volume is very small and running dozens of 1200W runs the background pressure is effected significantly by the chamber temperature which exceeded 100degC)
Voltage: 52-40kV (drop is based on increasing pressure)
Current: 24mA
TIER: 1,525,000 n/sec
1.5Mnsec-1200W-16Feb2019.png
I repeated this run approximately 30 times with results in the +/- 10% range.
User avatar
Rich Feldman
Posts: 1470
Joined: Mon Dec 21, 2009 6:59 pm
Real name: Rich Feldman
Location: Santa Clara County, CA, USA

Re: Joe Gayo's Fusor - Neutron Detection Upgrade

Post by Rich Feldman »

Nice. Please say more about how you set up the axial magnetic field -- or point to an earlier post if you already told the story. Permanent magnets? Steel yoke?
All models are wrong; some models are useful. -- George Box
User avatar
Joe Gayo
Posts: 404
Joined: Sun Jan 06, 2019 9:34 pm
Real name: Joe Gayo
Location: USA

Re: Joe Gayo's Fusor - Neutron Detection Upgrade

Post by Joe Gayo »

Qty. 2 of 5856K16 (from McMaster Carr) on one opposing pair of sides. The fields are aligned in the same direction. I used FEMM (http://www.femm.info/wiki/HomePage) to calculate the field strength.
User avatar
Joe Gayo
Posts: 404
Joined: Sun Jan 06, 2019 9:34 pm
Real name: Joe Gayo
Location: USA

Re: Joe Gayo's Fusor - Neutron Detection Upgrade

Post by Joe Gayo »

For the first few seconds, the count rate is 400/sec when the voltage is 50kV. As the pressure rises and the voltage falls, the count rate finishes at 200/sec. Roughly 100/sec correlates to 500k n/sec, so the rate starts at 2M n/sec and falls to 1M n/sec. If the pressure was dynamically controlled to compensate for the temperature rise, then 2M n/sec would be sustainable.