Yes, it starts at atmospheric pressure and then slowly drops. I generally wait until the thirty minutes is up to check the gauge after it has had sufficient time to heat up, but I'm pretty sure it gets down to near its lowest pressure immediatetly.

I had my eyes on this pump for sometime:
https://www.ebay.com/itm/EDWARDS-5-ROTA ... SwQS1aQn4L

It's a bit pricey but I'm sure I could negotiate.
The Youngest Fusioneer At 12 Years Old

User avatar
Richard Hull
Site Admin
Posts: 10809
Joined: Fri Jun 15, 2001 1:44 pm
Real name: Richard Hull

Re: A Bunch of Failures

Post by Richard Hull » Mon Jan 01, 2018 7:35 pm

That is a good price a working 5CFM Edwards.

Richard Hull
Progress may have been a good thing once, but it just went on too long. - Yogi Berra
Fusion is the energy of the future....and it always will be
Retired now...Doing only what I want and not what I should...every day is a saturday.

Jackson Oswalt
Posts: 97
Joined: Thu Jul 13, 2017 1:10 pm
Real name: Jackson Oswalt

Re: A Bunch of Failures

Post by Jackson Oswalt » Mon Jan 01, 2018 9:26 pm

Oddly enough, it's supposedly on 3.75 CFM, but is rated to reach 7.5 X 10^-4 Torr. I figure that will be enough to back a diffusion pump.
The Youngest Fusioneer At 12 Years Old

Rex Allers
Posts: 235
Joined: Sun Dec 30, 2012 8:39 am
Real name:
Location: San Jose CA

Re: A Bunch of Failures

Post by Rex Allers » Mon Jan 01, 2018 10:44 pm

Checking the specs, the 5 in E2M5 model is 5 m^3/hr pumping speed which I convert as about 3 cfm.
Rex Allers

User avatar
Richard Hull
Site Admin
Posts: 10809
Joined: Fri Jun 15, 2001 1:44 pm
Real name: Richard Hull

Re: A Bunch of Failures

Post by Richard Hull » Mon Jan 01, 2018 11:05 pm

Sorry, I forgot Edwards is Euro and thus SI. Still, it is good enough for a fusor, if in good shape.

Richard Hull
Progress may have been a good thing once, but it just went on too long. - Yogi Berra
Fusion is the energy of the future....and it always will be
Retired now...Doing only what I want and not what I should...every day is a saturday.

User avatar
Dennis P Brown
Posts: 1470
Joined: Sun May 20, 2012 2:46 pm
Real name: Dennis P Brown
Location: Glen Arm, MD

Re: A Bunch of Failures

Post by Dennis P Brown » Fri Jan 05, 2018 3:33 pm

That appears to be a good pump and will certainly work with any DP as long as it gets below 50 microns; so when you say
but is rated to reach 7.5 X 10^-4 Torr. I figure that will be enough to back a diffusion pump.
So, as long as it gets under 5 microns (assuming the system has no real leaks) you will be more than fine.

No mechanical pump (even new) will really do that sub-micron - I have a very large pump, a leak proof system (and back fill with dry air), and 2-3 microns is what it can do and that is excellent for most mechanical pumps.

User avatar
Bob Reite
Posts: 276
Joined: Mon Aug 26, 2013 1:03 am
Real name: Bob Reite
Location: Wilkes Barre/Scranton area

Re: A Bunch of Failures

Post by Bob Reite » Sat Jan 06, 2018 3:16 am

IMHO, a mechanical pump is doing well to get into the single digit micron range.
The more reactive the materials, the more spectacular the failures.
The testing isn't over until the prototype is destroyed.

Jackson Oswalt
Posts: 97
Joined: Thu Jul 13, 2017 1:10 pm
Real name: Jackson Oswalt

Re: A Bunch of Failures

Post by Jackson Oswalt » Sat Jan 06, 2018 4:02 am

I knew not to expect it to reach that pressure. I learned not to expect things like that from my HVAC pump. Anyways, the Edwards pump is here now and, of course, I can't find anyone selling adapters from 1 bsp to kf25 that doesn't ship from Hong Kong.
The Youngest Fusioneer At 12 Years Old

John Futter
Posts: 1289
Joined: Thu Apr 22, 2004 2:29 am
Real name:
Contact:

Re: A Bunch of Failures

Post by John Futter » Sat Jan 06, 2018 7:35 am

Anyone with a lathe some brass tube and a little brass plate would make one in under an hour
no oxy acetylene no prob just soft solder will do.

MAKE ONE!!!!

User avatar
Richard Hull
Site Admin
Posts: 10809
Joined: Fri Jun 15, 2001 1:44 pm
Real name: Richard Hull

Re: A Bunch of Failures

Post by Richard Hull » Sat Jan 06, 2018 8:36 pm

Bob and John are correct.

A used pump is doing fabulous if it easily hits 15 microns within a minute or two at the head. If it goes below 10 microns, it was well cared for or not used much in a good environment.

Where there is a will there is a way. KF Adapters can be whipped out quickly by even a novice machinist. Made in brass, they are easily soft soldered.

I have had to do this a number of times back in the 1990's as I used a lot of HVAC pumps. Some good and some not so good. The worst ever did hit 30 microns rapidly. The best was a 5 CFM yellow jacket that hit 12 microns at the head in under 40 seconds! Now, I use only belt drive pumps which on whole are no better or worse in evacuating a system but are much, much more quiet and easy on the nerves, easy to repair and replace motors.