NT150/360 Problem
-
- Posts: 405
- Joined: Wed May 08, 2013 3:36 pm
- Real name: Tom McCarthy
- Location: Ireland
- Contact:
NT150/360 Problem
I tried running my turbopump again today after a hiatus of a couple of months.
When I started the turbopump (press "Start" on the controller) it started accelerating, and then hit normal operation. Soon after hitting normal operation, the red failure light lit up and I pressed stop on the controller, shut off the backing pump and let some air in via the foreline to slow things down. I gave the pump a few minutes and pumped down again. On reaching 20 microns again, I turned the controller on. This time after the controller turns on and is finished self-testing, the failure light is lit up, without starting the turbo at all. Disconnecting the controller from mains power and connecting again results in the same.
It's a Leybold Oerlikon TMP 150, with a NT 150/360 controller. The pump was serviced and certified fine by Leybold last year, and has not been used since, same for the controller.
Backing pressure was 20 microns or so. Backing pump is an Edwards RV3, I've posted about both before. Leaks are negligible, it takes over 3 minutes to go from 20 to 40 microns when the valve on the roughing pump is shut.
John Futter seems to be a fan of the TMP pumps, but is not so fond of their NT controllers - Any ideas John?
When I started the turbopump (press "Start" on the controller) it started accelerating, and then hit normal operation. Soon after hitting normal operation, the red failure light lit up and I pressed stop on the controller, shut off the backing pump and let some air in via the foreline to slow things down. I gave the pump a few minutes and pumped down again. On reaching 20 microns again, I turned the controller on. This time after the controller turns on and is finished self-testing, the failure light is lit up, without starting the turbo at all. Disconnecting the controller from mains power and connecting again results in the same.
It's a Leybold Oerlikon TMP 150, with a NT 150/360 controller. The pump was serviced and certified fine by Leybold last year, and has not been used since, same for the controller.
Backing pressure was 20 microns or so. Backing pump is an Edwards RV3, I've posted about both before. Leaks are negligible, it takes over 3 minutes to go from 20 to 40 microns when the valve on the roughing pump is shut.
John Futter seems to be a fan of the TMP pumps, but is not so fond of their NT controllers - Any ideas John?
- Richard Hull
- Moderator
- Posts: 14991
- Joined: Fri Jun 15, 2001 9:44 am
- Real name: Richard Hull
Re: NT150/360 Problem
Just curious... Is that pump spec'd to run in the horizontal position?
Richard Hull
Richard Hull
Progress may have been a good thing once, but it just went on too long. - Yogi Berra
Fusion is the energy of the future....and it always will be
The more complex the idea put forward by the poor amateur, the more likely it will never see embodiment
Fusion is the energy of the future....and it always will be
The more complex the idea put forward by the poor amateur, the more likely it will never see embodiment
-
- Posts: 405
- Joined: Wed May 08, 2013 3:36 pm
- Real name: Tom McCarthy
- Location: Ireland
- Contact:
Re: NT150/360 Problem
Yeah, Leybold give the go-ahead for horizontal operation. That's not the bug unfortunately.
From the manual:
In most applications the TURBOVAC will be flanged direct to the high-vacuum
flange for the system. The design of the lubricating system makes it
possible to mount and run the TURBOVAC in any desired attitude. It is not
necessary to support the pump.
From the manual:
In most applications the TURBOVAC will be flanged direct to the high-vacuum
flange for the system. The design of the lubricating system makes it
possible to mount and run the TURBOVAC in any desired attitude. It is not
necessary to support the pump.
-
- Posts: 975
- Joined: Tue Nov 24, 2009 3:08 am
- Real name:
- Location: Beaverton, OR
Re: NT150/360 Problem
Do NOT let air in the foreline to slow it down. Some pumps have a orifice on a solenoid that can be opened at around half speed. This is not one of them.
Probably something in the controller died. It is a pretty old model so the caps could be going south. You can call olerikon and talk to them and maybe they will send you a schematic like they did for my NT50 and then you can start chasing down the fault. Call them. Don't email them.
But a warning, these drives are not isolated from the mains so all your test equipment will need to be or you need to put the drive on an isolation transformer.
Probably something in the controller died. It is a pretty old model so the caps could be going south. You can call olerikon and talk to them and maybe they will send you a schematic like they did for my NT50 and then you can start chasing down the fault. Call them. Don't email them.
But a warning, these drives are not isolated from the mains so all your test equipment will need to be or you need to put the drive on an isolation transformer.
-
- Posts: 1848
- Joined: Wed Apr 21, 2004 10:29 pm
- Real name: John Futter
- Contact:
Re: NT150/360 Problem
been away a few days
I have a full manual for that controller but as jerry says it is very old
there is a newer controller bigger in size that runs all these pumps from your one up to the tp1000C
can nhot rmrmber the part # will look at work after Easter as we have three of thesec all off ebay.
Check the cable that controller uses a couple of the wires to look at the pump for an internal resistor that tells it whether it is a 150 or 300 l/s version
(differnt speeds and accel ramp) a borken wire in the cable will bring up the fault light
but it sounds like the half bridge output stage has a shorted TO3 trnasistor in it
I have a full manual for that controller but as jerry says it is very old
there is a newer controller bigger in size that runs all these pumps from your one up to the tp1000C
can nhot rmrmber the part # will look at work after Easter as we have three of thesec all off ebay.
Check the cable that controller uses a couple of the wires to look at the pump for an internal resistor that tells it whether it is a 150 or 300 l/s version
(differnt speeds and accel ramp) a borken wire in the cable will bring up the fault light
but it sounds like the half bridge output stage has a shorted TO3 trnasistor in it
-
- Posts: 405
- Joined: Wed May 08, 2013 3:36 pm
- Real name: Tom McCarthy
- Location: Ireland
- Contact:
Re: NT150/360 Problem
Thanks for calling that Jerry, what would you do when wanting to slow down the pump then - wait it out? There's no real speed feedback on the controller, but the manual says it's able to start the pump even if it's still rotating, maybe I don't need to slow the pump down.
John, will be waiting to hear back from you. Cable, pump and controller were all provided by a reputable and generous vacuum business here in Ireland https://www.provac.ie/ and pump was working previously so cable was okay too.
John, will be waiting to hear back from you. Cable, pump and controller were all provided by a reputable and generous vacuum business here in Ireland https://www.provac.ie/ and pump was working previously so cable was okay too.
-
- Posts: 975
- Joined: Tue Nov 24, 2009 3:08 am
- Real name:
- Location: Beaverton, OR
Re: NT150/360 Problem
You wait. Pumps can be restarted it will apply a bit of power and look at the back emf and determine the speed and start ramping up from there.
Waiting a few minutes is better than letting in too much air and turning your pump into a bunch of shredded aluminum foil. It's also hell on the bearings.
I would call up the supplier. They might be able to help, maybe they have another drive around you could at least borrow to narrow the problem down to the pump or drive.
Waiting a few minutes is better than letting in too much air and turning your pump into a bunch of shredded aluminum foil. It's also hell on the bearings.
I would call up the supplier. They might be able to help, maybe they have another drive around you could at least borrow to narrow the problem down to the pump or drive.
-
- Posts: 405
- Joined: Wed May 08, 2013 3:36 pm
- Real name: Tom McCarthy
- Location: Ireland
- Contact:
Re: NT150/360 Problem
Suppliers are going taking a look at it and see what they can do, sent it off today.
-
- Posts: 1848
- Joined: Wed Apr 21, 2004 10:29 pm
- Real name: John Futter
- Contact:
Re: NT150/360 Problem
Tom
Start looking for a NT20 controller a much later controller for most of the leybold pumps
heres the manual for it
Start looking for a NT20 controller a much later controller for most of the leybold pumps
heres the manual for it
-
- Posts: 405
- Joined: Wed May 08, 2013 3:36 pm
- Real name: Tom McCarthy
- Location: Ireland
- Contact:
Re: NT150/360 Problem
Vacuum company fixed the controller thankfully. John, think you were right, it was a power transistor.