Portable XRF detector window repair

This area is for discussions involving any fusion related radiation metrology issues. Neutrons are the key signature of fusion, but other radiations are of interest to the amateur fusioneer as well.
Post Reply
steve_rb
Posts: 145
Joined: Sun Oct 18, 2009 4:18 am
Real name: Steve Robinson
Location: Tehran University

Portable XRF detector window repair

Post by steve_rb »

After droping Be window of our portable XRF detector is ruptured. Detector is very expensive and I am trying to fix the window. Be window thickness is about 13 micron (wow too thin). Thin papers are about 100 micron in comparision. I can purchase 13 micron custom ordered Be windows and do soldering after heating and removing the reminder of the ruptured Be window inside a vacuum chamber but I am thinking about using Hysol resin instead of soldering. Also I am thinking about using 50-100 micron Kapton window instead of Be window and use hysol resin to stick Kapton window in place. Any suggestion, recommendation or advice in this regard?
Attachments
detector4.jpg
User avatar
Dennis P Brown
Posts: 3159
Joined: Sun May 20, 2012 10:46 am
Real name: Dennis Brown

Re: Portable XRF detector window repair

Post by Dennis P Brown »

Don't know your application but if kapton can be used instead that would be a safer method. Kapton is very strong and depending on whether that is a pressure window or vacuum, the film could be thinner which might be useful. Or maybe not - again, depends on its application and enviroment. Beryllium is toxic and when it breaks it creates dust particles, too. Never vacuum up the particles except with a HEPA filtered vacuum. Disposal is, of course, an issue since it is a hazardous waste and must be carefully packed, labled and disposed of properly. That is why we gave up on Beryllium (its very brittle nature tends to enhance breakage.)
User avatar
Maciek Szymanski
Posts: 193
Joined: Wed Nov 20, 2019 2:31 pm
Real name: Maciek Szymański
Location: Warsaw, Poland

Re: Portable XRF detector window repair

Post by Maciek Szymanski »

Standard practice at my laboratory for mounting thin foils (down to 1.8μ) is to use either minute amount of epoxy or shellac diluted in alcohol. We use a standard epoxy (even not a low vapor pressure one), sometimes the conducting one. A very thin layer of glue is usually applied with a whisker tool under a dissecting microscope. No problems with vacuum compatibility.
“Begin at the beginning," the King said, very gravely, "and go on till you come to the end: then stop.” ― Lewis Carroll, Alice in Wonderland
steve_rb
Posts: 145
Joined: Sun Oct 18, 2009 4:18 am
Real name: Steve Robinson
Location: Tehran University

Re: Portable XRF detector window repair

Post by steve_rb »

Application is detecting low energy X-rays. I found two roll of Mylar films (12.5 and 25 micron thick) inside my old stuff which seems are made just for low energy x-ray transmission applications. I think I can use 12.5 micron one which looks the best for my case. The only think left is whether or not I need to remove the left Be ring (see above photo) around the detector opening. It shouldn't be hard. Just by heating probably up to 200 DegC solder will liquidized and Be will come off (I think of course, never done this before). If I remove Be epoxy will sit on the solder but if I don't remove left Be epoxy will sit on the Be. Which one will is better and will cause Mylar film to stick to it strongly I don't know. Any other advice?

Another problem is visible light which will go into the detector which should be prevented. Be obstruct the visible light but Mylar film will not. I think I should get Mylar film to be Al coated a few angstrom or I have to black paint it in top side with a black ink or something like that. Any advice in this?

Also there is another problem. After applying resin and starting evacuation at the end when vacuum reached 10^-6 tor I will push the film towards detector and hold it there (using a spring behind the film frame for about 48 hours for the epoxy to be cured). From this point all out gassing will be trapped inside the detector and will increase pressure inside the detector and can affect detector performance. So I need a zero out gassing epoxy. Any suggestion which epoxy has zero outgassing for this purpose?
User avatar
Dennis P Brown
Posts: 3159
Joined: Sun May 20, 2012 10:46 am
Real name: Dennis Brown

Re: Portable XRF detector window repair

Post by Dennis P Brown »

The x-rays if very soft might have issues with 12 micron thick plastic. Beryllium is safe to handle if no dust is created so that might still be your best bet. Broken Be is a hazard and needs to be addressed as such. It can't be thrown into normal trash.

Very small amounts of epoxy aren't a big issue with 10^-6 torr (through not below) but I'd use proper vacuum grade if available. Certainly use as little as possible. A Beryllium surface should be fine for epoxy. Paint could cause issues with x-rays so maybe not a good idea (the Ti and other elements can absorb.) The Al coating would likely need to be over 600 A to stop most (not all) light. Again, that might be an issue for soft x-rays - depends on what you are trying to measure.
John Futter
Posts: 1848
Joined: Wed Apr 21, 2004 10:29 pm
Real name: John Futter
Contact:

Re: Portable XRF detector window repair

Post by John Futter »

use new Be foil
http://www.goodfellow.com/catalogue/GFC ... V2hvddblfZ
BE000210 looks right.
Be careful it is not good for you
you need the Be foil to keep it light tight and to remove the low energy spectrum (noise)
steve_rb
Posts: 145
Joined: Sun Oct 18, 2009 4:18 am
Real name: Steve Robinson
Location: Tehran University

Re: Portable XRF detector window repair

Post by steve_rb »

John Futter wrote: Wed Oct 21, 2020 2:02 pm use new Be foil
http://www.goodfellow.com/catalogue/GFC ... V2hvddblfZ
BE000210 looks right.
Be careful it is not good for you
you need the Be foil to keep it light tight and to remove the low energy spectrum (noise)
BE000210 diameter is 100 mm. I need 120 mm.

I have a great idea would like to share and have comments. I was worried about brazing Be or gluing Mylar inside the vacuum chamber but I though I can do this outside. I mean drill a 1 mm hole at the side of the detector first and then do Be brazing or Mylar gluing and curing. Now put the window fixed detector inside the vacuum chamber and start evacuation until pressure go down to 10^-6 and then solder block that 1 mm hole at the side of the detector I made before inside the vacuum. Solder blocking this 1 mm hole inside the vacuum is lot easier that soldering Be window or gluing and curing Mylar inside the vacuum. Beside if there is any degassing during Mylar curing will go out during evacuation.
John Futter
Posts: 1848
Joined: Wed Apr 21, 2004 10:29 pm
Real name: John Futter
Contact:

Re: Portable XRF detector window repair

Post by John Futter »

I do not understand
I put up 15micron thickness solution over your 13 micron measurement
and you need 120mm when you said originally 13 microns

or is the detector nearly 5 inches in diameter --i doubt\\

please clarify!!!!
User avatar
Dennis P Brown
Posts: 3159
Joined: Sun May 20, 2012 10:46 am
Real name: Dennis Brown

Re: Portable XRF detector window repair

Post by Dennis P Brown »

I too am a bit confused; why braze the Be window when it could be epoxied just as well.
Andrew Seltzman
Posts: 815
Joined: Sun Feb 01, 2004 8:02 pm
Real name: Andrew Seltzman
Contact:

Re: Portable XRF detector window repair

Post by Andrew Seltzman »

I can confirm that hysol-1C epoxied Be windows will work down below 1e-6 torr. I was using an in-house build windows on my PhD research, they worked quite well.
The window on the the detector prevents condensation wince the detector is TEC cooled, it may need to be evacuated, though it could work at atmospheric pressure with dry air potentially with greater heat load on the cooler for the detector.
Andrew Seltzman
www.rtftechnologies.org
steve_rb
Posts: 145
Joined: Sun Oct 18, 2009 4:18 am
Real name: Steve Robinson
Location: Tehran University

Re: Portable XRF detector window repair

Post by steve_rb »

Sorry my stupid mistake. Detector diameter is 12 mm.
Detector is a SiPN type and electronically cools down to about -27 DegC that is why I should evacuate inside the detector casing otherwise detector tempreature will do down up to -15 DegC and device will not start working.
Epoxy is good option but solder option also is not very difficult to do (specially now that it will be done outside the vacuum) and I think is better than epoxy but spoxy is the must if soldering was not successful for any reason.

Also Be is preferred but if I couldn't find 12.5 micron thickness and 12 mm diameter Be foil I can use 12.5 micron Mylar I have and get it coated with Al in one or both sides. Mylar is easy to cut to 12 mm diameter but Be foil is not easy to cut.
John Futter
Posts: 1848
Joined: Wed Apr 21, 2004 10:29 pm
Real name: John Futter
Contact:

Re: Portable XRF detector window repair

Post by John Futter »

I use a wad cutter to cut foils
Be is used as it is the first stable (sort of) metal in the periodic table ie low Z so low energy gammas / x-rays can go throught it
Nacim maamri
Posts: 2
Joined: Fri Feb 23, 2024 2:51 pm
Real name: Nacim maamri

Re: Portable XRF detector window repair

Post by Nacim maamri »

Hi steve !
I have the same problem as you.
I have a spectrometer detector with the beryllium be window broken I'm a little lost.
Do you have addresses to recommend to me to find or buy beryllium foil?
How did you manage to glue the beryllium foil